发布时间:2019-04-23 12:29 原文链接: 3RACEPCR

实验概要

 

        This technique is used to obtain the 3'end of a cDNA, it requires some sequence information internal to the mRNA under study. The sequence information obtained from this technique can be utilised to obtain full length cDNA clones using the 5'RACE technique.
        The method I've used is loosely based on that given in 'PCR protocols: A guide to methods and applications', Academic Press Inc.
        The protocol given uses SuperScript II reverse transcriptase (Life Technologies) and Taq polymerase from Boehringer Mannheim. You will need to adjust the reagents according to the enzymes you use.

 


 

主要试剂

    Sterile water (high purity, RNase-free) 

    5 x SuperScript RTase buffer (Life Technologies) 

    100mM DTT 

    2.5mM dNTPs 

    RNase inhibitor (Pharmacia) 

    SuperScript RTase (Life Technologies) 

    10 x PCR buffer (Boehringer Mannheim, contains 15mM MgCl2) 

    500mM dNTPs 

    Mineral oil 

    RNA : Both total and polyA RNA are suitable for this technique. I recently compared the two and found the polyA RNAreaction to have the edge, but  
                only just. I do recommend using polyA RNA for 5'RACE though. 

    Primers: 

                T17 Adapter primer (T17AP): GACTCGAGTCGACATCGATTTTTTTTTTTTTTTTT 

                Adapter primer (AP): GACTCGAGTCGACATCG 

                Gene specific primer: Designed from known sequence, ideally it should an 18-22-mer with an annealing temperature of around 56°C, where
                 a G/C = 4°C and an A/T = 2°C.

 

实验步骤

1. Place approx. 1ug of total RNA (~10ng polyA RNA) in an RNase-free eppendorf, make up to 10ul with pure water. Add 1ul of T17AP (0.5ug/ul).

 

2. Heat this to 70°C for 10 minutes, then place on ice for 5 minutes.

 

3. Spin pulse the tube and then add the following reagents:

 

    4ul 5X SuperScript RTase buffer 

    2ul 100mM DTT 

    1ul 2.5mM dNTPs 

    1ul RNase inhibitor 

    1ul SuperScript RTase.

 

4. Mix and incubate at 37°C for 2 hours.

 

5. Use a serial dilution of the reverse transcription reaction: 2ul of undiluted cDNA, 2ul of 10 x diluted cDNA and 2ul of 100 x diluted cDNA in the PCR reaction. Set up the PCR as follows:

    35.5ul pure water 

    5ul 10 x PCR buffer (Boehringer Mannheim) 

    5ul 500uM dNTPs (some enzymes require a higher concentration of dNTPs (2mM), for example Pfu polymerase (though I'd recommend against the  
       use of this enzyme) 

    1ul Adapter primer (25uM) 

    1ul Gene specific primer (25uM) 

    2ul cDNA template (various dilutions)

 

6. Mix and overlay with mineral oil.

 

7. Perform the following PCR cycles:

    a. 94°C 3 minutes 

    b. 72°C 3 minutes (at this point add 0.5ul of Taq polymerase) 

    c. 94°C 30 seconds 

    d. 56°C* 1 minute 

    e. 72°C 2 minutes (longer if expected fragment size is greater than 2kb) 

    f. Cycle steps c. - e. 35 times 

    g. 72°C 10 minutes.

 

* Varies according to the melting temperature of your gene specific primer. The annealing temperature in the PCR may require some adjustment for optimal yield of PCR product.

 

8. Load all or part of the reaction on an agarose gel. Don't be too surprised if you get a faint background smear as well as your prominent band


相关文章

罗氏诊断首款国产PCR一体机获批

1月5日,国家药监局发布医疗器械批准证明文件,其中罗氏诊断产品(苏州)有限公司最新获批一款全自动核酸提取及荧光PCR检测系统(国械注准20253222730)。这是目前罗氏诊断获批的首个国产PCR一体......

4500万+预算,海关总署75台套仪器设备重新招标

近日,海关总署物资装备采购中心发布多则公告,分别就“海关总署2025年数字化摄影X射线系统采购项目(重新招标)”“海关总署2025年PCR仪采购项目(重新招标)”“海关总署2025年彩色多普勒超声诊断......

闪量科技全自动PCR分析系统列入国家重点研发计划的医疗器械

医疗器械优先审批申请审核结果公示(2025年第14号)依据原国家食品药品监督管理总局《医疗器械优先审批程序》(总局公告2016年168号),对申请优先审批的医疗器械注册申请进行审核,现将符合优先审批情......

超6900万元!海关总署最新一批液质、气相、PCR等仪器中标出炉

近日,海关总署陆续公布了一批分析仪器的采购项目中标结果,涵盖气相色谱仪、超高效液相色谱-三重四极杆质谱仪、波长色散X射线荧光光谱、PCR仪蛋白质测定仪、基因测序仪等多个领域,总中标金额高达6900余万......

伯乐公司通过战略收购与平台推出扩展四款数字PCR产品系列

全球生命科学研究和临床诊断产品领域的领导者伯乐实验室有限公司(纽约证券交易所代码:BIO和BIO.B)近日宣布推出四款新的微滴式数字PCR(ddPCR™)平台。新推出的仪器包括伯乐公司的QXConti......

医疗器械优先审批申请审核结果公示(2025年第6号)

国家药品监督管理局医疗器械技术审评中心发布医疗器械优先审批申请审核结果公示(2025年第6号),同意了苏州淦江生物技术有限公司申请的运动神经元存活基因1(SMN1)检测试剂盒(PCR-荧光探针熔解曲线......

3280万元华中农业大学大批仪器采购意向涉PCR、蛋白纯化等

近日,华中农业大学发布多个实验室仪器设备政府采购意向,采购的产品包括:超高分辨多色快速成像系统、荧光定量PCR仪、显微镜、蛋白纯化系统、分析天平、电泳仪、摇床、细胞破碎仪、核酸转染系统、纯水系统等,采......

伯乐领投Geneoscopy公司完成1.05亿美元的C轮融资

Geneoscopy公司周三宣布已完成1.05亿美元的C轮融资。此轮融资由伯乐实验室领投,两家公司在一份联合声明中表示,这笔资金将用于支持Geneoscopy公司无创结直肠癌筛查检测的推出。参与此次融......

9.5个亿!伯乐收购PCR公司

波兰生命科学公司ScopeFluidics近日表示,在收到交易的最后一笔款项后,该公司最近敲定了以1.3亿美元(约合9.5亿元人民币)的价格将其子公司CuriosityDiagnostics出售给Bi......

北京市聚合酶链反应(PCR)检验实验室检查指南(2024版)

京药监发〔2024〕261号各区市场监管局,房山区燕山市场监管分局,市市场监管局机场分局,经开区商务金融局,市药监局各分局,各相关事业单位:为深入贯彻落实医疗器械生产监管相关法规要求,进一步规范北京市......