实验概要

The procedure  presented below describes a facile method for studying signal  transduction events with adherent cells (HeLa, MCF-7, BALB/c 3T3, etc.).  In this procedure, cells are gently lifted from their culture vessel,  placed into wells of a 96-well filter bottom plate, and stimulated as  desired. At the end of stimulation, cell culture medium is removed from  the bottom of the wells by gentle aspiration using a vacuum manifold.  The cells are then washed with PBS, aspirated, and lysed within the  wells by addition of Cell Extraction Buffer. The cell extracts are then  assayed using Invitrogen™ p38MAPK [pTpY180/182] and p38MAPK (total)  phosphoELISA™ kits.

实验材料

The following materials will be needed:

  • Trypsin/EDTA solution (Cat. no. 25300-054)

  • Complete Cell Culture Medium (e.g., DMEM plus 10% FBS)

  • PBS (Invitrogen Cat. nos.11885-076 and 26140-079)

  • Cell Extraction Buffer (Cat. no. FNN0011)

  • Orbital Plate Shaker (Lab Line Titer Plate Shaker Cat. no. 4625-EA)

  • p38 MAPK [pTpY180/182] phosphoELISA™ (Cat. no. KHO0061)

  • p38 MAPK Total phosphoELISA™ (Cat. no. KHO0071)

  • BD Falcon 96-well cell culture plate (BD Cat. no. 353072)

Cell extraction buffer formulation:

  • 10 mM Tris, pH 7.4 • 2 mM Na3VO4

  • 100 mM NaCl • 1% Triton X-100

  • 1 mM EDTA • 10% glycerol

  • 1 mM EGTA • 0.1% SDS

  • 1 mM NaF • 0.5% deoxycholate

  • 20 mM Na4P2O7 • 1 mM PMSF (stock 0.3 M in DMSO)

Prepare Protease Inhibitor Cocktail (Sigma Cat. no. P-2714) according  to the manufacturer’s guideline as a 10x stock. Add 100 μl per 1 ml  Cell Extraction Buffer.

实验步骤

1.      Grow cells to desired level of confluency in a T75 flask.

2.      Decant or aspirate the medium.

3.      Add 2–3 ml fresh warm trypsin/EDTA solution. Transfer the flask to a 37°C incubator.

4.      Wash with warm PBS. Aspirate.

5.       After 5 minutes, tap the side of the flask, and examine the flask under  a microscope for lifting. If necessary, return the cells to the  incubator for an additional 5–10 minutes, with occasional tapping, until  lifting is complete.

6.      Quickly quench the Trypsin reaction by adding 5–6 ml Complete Cell Culture Medium.

7.      Transfer the cells to sterile 15 ml conical tubes.

8.      Pellet the cells by centrifugation at 300 x g for 7 minutes.

9.      Decant the supernatant.

10.   Wash the cells by pipetting 10 ml medium into each conical tube and  resuspending the pellet. Collect the cells by centrifugation at 300 x g  for 7 minutes.

11.  Resuspend the washed cells in complete cell culture medium.

12.  Enumerate cell density. For most applications, the cell density should be adjusted to 5–25 x 104  cells/ml cell culture medium. It is important to note that this value  may require some optimization for each specific application. Cell  doubling time is an important factor to be considered when adjusting  cell density at the beginning of an experiment.

13.   Plate 200 μl of cell culture (i.e., 10,000–50,000 cells) into the wells  of the sterile 96-well cell culture plate. Incubate the cells for 18  hours at 37°C.

14.   Stimulate the cells as desired. In the example presented below, HeLa  cells were treated with 100 μM anisomycin for 60 minutes at 37°C.

15.  At the end of the treatment, aspirate the media from the wells.

16.   Wash the cells by pipetting 200 μl ice-cold PBS into each well. Remove  the PBS by aspiration and repeat two times for a total of three  washings.

17.   Aspirate the medium. Pipette 25 μl protease-inhibitor-supplemented Cell  Extraction Buffer into each well. Incubate the plate on ice for 30  minutes.

18.   Thoroughly mix the contents of each well by pipetting up and down 5–6  times. A multi-channel pipette is desirable for this application. This  procedure causes the cells to lyse. At this point in the procedure, the  extracts are ready for analysis. Alternatively, the extracts may be  stored in the filter bottom plate at –20°C for future analysis. Frozen  plates should be thawed on ice in preparation of completing the assays.

19.  Place the plate on an orbital shaker and mix for 1 minute.

20.   Prepare the phosphoELISA™ kits. Sample Wells: Pipette 95 μl Standard  Diluent Buffer (included in the kits) into the wells of the  phosphoELISA™ plates designated for samples. Transfer 5 μl cell extract  from the filter plate into the sample wells of the plates. Place the  plates on an orbital shaker to thoroughly mix the contents of the wells.  Standard Wells: Prepare standards as indicated in the assay protocol  and pipette into designated wells.

21.  Complete the phosphoELISA™ as directed by the assay protocol.