发布时间:2019-04-23 16:47 原文链接: 3ColorStaining:AlphaCatenininHeLaHumanCervicalCancerCells

实验概要

Alpha-catenin  in HeLa human cervical cancer cells was labeled using mouse  anti-α-catenin and visualized with Alexa Fluor® 488 goat anti-mouse IgG  (green). Filamentous actin was visualized using red-fluorescent Alexa  Fluor® 635 phalloidin. Nuclear DNA was stained with blue-fluorescent  DAPI.

主要试剂

 

10X Phosphate-buffered saline (PBS) [详细信息:GIBCO®]

Triton X-100 [详细信息:Sigma]

Methanol-free 16% formaldehyde solution [详细信息:Polysciences]

Bovine serum albumin [详细信息:Sigma]

Normal goat serum (10%) [详细信息:Invitrogen Cat. no. 50-19Z]

Anti-α-catenin antibody, mouse monoclonal [详细信息:Invitrogen Cat. no 180225]

Alexa Fluor® 488 goat anti-mouse IgG

Alexa Fluor® 635 phalloidin

DAPI

Prolong® Gold antifade reagent

 

实验材料

 

Confluent HeLa human cervical cancer cells

实验步骤

 

(All steps are carried out at room temperature unless otherwise noted.)

1. Sample Preparation

1) Remove media from live cells and fix cells for 15 minutes in 4% formaldehyde solution.

2) Wash cells 3 x 5 minutes in PBS (1X).

3) Permeabilize cells by treating for 10 minutes with PBT (0.1% Triton X-100 in PBS).

4) Wash cells 3 x 5 minutes in PBS.

2. Cell Staining

1) Block cells for 1 hour in antibody blocking buffer (10% normal goat serum (NGS), 1% BSA in PBT).

2) Incubate cells with 5 µg/ml of anti--catenin primary antibody in antibody blocking buffer for 45 minutes.

3) Wash cells 3 x 10 minutes in PBT.

4) Incubate cells with 5 g/ml Alexa Fluor® 488 goat anti-mouse IgG in 1% BSA/PBT for 30 minutes.

5) Wash 3 x 5 minutes in PBS.

6) Incubate cells for 20 minutes with Alexa Fluor® 635 phalloidin, diluted 1:40 from stock solution.

7) Wash 3 x 5 minutes in PBS.

8) Mount the specimen in ProLong® Gold antifade reagent.

3. Mounting Reagent Preparation and Sample Processing

1) Remove the ProLong Gold antifade reagent from the freezer and allow the vial to equilibrate to room temperature.  NOTE:   Using an external heat source to warm the vial is not recommended, as this may decrease the long-term stability of the product.

2) Remove  any excess liquid from the specimen and apply 1 or 2 drops (depending  on the surface area of your sample) of the antifade reagent to the  specimen.

3) Cover  slide-mounted specimens with a coverslip; for specimens mounted on  coverslips, place a drop of antifade reagent onto a clean slide and  carefully lower the coverslip onto the antifade reagent to avoid  trapping any air bubbles.

4) Allow  the mounted sample to cure on a flat surface in the dark. Curing time  may vary from a couple of hours to overnight, depending on the thickness  of the sample and the relative humidity of the surrounding air.

5) For  long-term storage, seal the coverslip to the slide after curing to  prevent excessive shrinkage of the mounting medium, which can result in  sample distortion. After sealing, store the slide upright in a covered  slide box at <–20°C. Desiccant may be added to the box to ensure that  the slide remains dry.

6) To  view the samples immediately, secure the coverslip at the corners using  nail polish or hot wax to prevent the coverslip from moving. Leave the  edges clear to allow the preparation to cure.

 

注意事项

 

1)         The antifade properties of ProLong® Gold antifade reagent improve slightly the longer it remains in contact with the specimen.

2)         To  further reduce photobleaching, minimize the exposure of fluorescently  labeled specimens to light by using neutral density filters, and expose  samples only when observing or recording a signal. Optimize image  capture by using a minimum of optics, high-numerical aperture  objectives, relatively low magnification, high-quality optical filters,  and high-speed film or high-efficiency detectors.

 

附    件   (共1个附件,占199KB)

3色染Cells_jpg.g

199KB  查看


相关文章

科研家提出环形RNA全长转录本解析技术

近日,中国科学院北京生命科学研究院赵方庆团队在《自然-实验手册》(NatureProtocols)上,发表了题为Full-lengthcircularRNAprofilingbynanoporeseq......

5分钟了解hela细胞形态

HeLa细胞通过关闭某些信号通路(包括PI3K和MKK4/MAPK信号通路),激活某些信号通路(包括cAMP、Ca~(2+)—CaM、PKC、Ras和P38/MAPK信号通路),激活或抑制不同的信号分......

原因几何?数十万篇科研文章结果竟难以重复

生物学和技术变异可以影响研究的可重复性,但大多数重点是去除后者。近期,耶鲁大学医学院Aebersold及刘延盛共同通讯在NatureBiotechnology在线发表题为“Multi-omicmeas......

不同实验室的HeLa细胞在遗传学和表型上存在异质性

HeLa细胞现已在世界各地的许多实验室中培养了将近70年,并且长期以来被认为是无限供应的没有发生变化的相同细胞。然而,在一项新的研究中,来自瑞士苏黎世联邦理工学院、日内瓦大学、巴塞尔大学、中国科学院、......

多组学技术发现13个实验室的HeLa细胞系存在显著变异

生物学研究的重复性一直是研究者重点关注的问题,除去实验误差、统计学P值、数据采集质量、研究者主观因素等,实验材料不同也可能造成迥异的实验结果。肿瘤细胞系作为一种重要实验材料,在各研究中被广泛使用。然而......

一种新的中心体蛋白质TACP1与有丝分裂激酶Nek2A相互作用的研究

[摘要]目的:研究一种新的中心体蛋白TACP1蛋白与有丝分裂激酶Nek2A在有丝分裂期的相互作用关系。方法:在大肠杆菌中表达和纯化的Nek2A3052446蛋白与293T细胞中表达的TACP1蛋白,进......

顺铂作用HeLa细胞的蛋白质组学研究

【摘要】目的:用双向电泳———质谱法分析顺铂作用于人宫颈癌HeLa细胞后蛋白质组的变化。方法:体外培养HeLa细胞,用MTT法观察不同时间、不同浓度顺铂对细胞的增殖抑制作用。加入10μg/mL顺铂作用......

HeLa细胞蛋白质组双向电泳技术的建立

【摘要】目的:建立和优化HeLa细胞蛋白质组分析所需的样品处理方法和双向电泳技术。方法:用三种不同配方的裂解液提取HeLa细胞中的蛋白质,进行双向凝胶电泳,用AmershamImageMasterVD......