实验概要

The  procedure presented below describes a facile method for studying signal  transduction events with suspension cells (Jurkat, Raji, THP-1, etc.).  In this procedure, cells are plated into the wells of 96-well filter  bottom plates and stimulated as desired. At the end of the stimulation,  cell culture medium is removed from the bottom of the wells by gentle  aspiration using a vacuum manifold. The cells are then washed with PBS,  aspirated, and lysed within the wells by addition of Cell Extraction  Buffer. The cell extracts are then assayed using Invitrogen™ p38MAPK  [pTpY180/182] and p38MAPK (total) phosphoELISA™ kits.

主要试剂

 

PBS, ice-cold [详细信息:Cat. no. 70013-032]

Cell Extraction Buffer [详细信息:Cat. no. FNN0001]

p38 MAPK [pTpY180/182] phosphoELISA™ [详细信息:Cat. no. KHO0061]

p38 MAPK Total phosphoELISA™ [详细信息:Cat. no. KHO0071]

Cell extraction buffer formulation:

10 mM Tris, pH 7.4

2 mM Na3VO4

100 mM NaCl

1% Triton X-100

1 mM EDTA

10% glycerol

1 mM EGTA

0.1% SDS

1 mM NaF

0.5% deoxycholate

20 mM Na4P2O7

1 mM PMSF (stock 0.3 M in DMSO)

 

主要设备

 

Orbital Plate Shaker [详细信息:Lab Line Titer Plate Shaker Cat. no. 4625-EA]

实验材料

 

Sterile 96-well filter bottom plates, 1.2 μm pore size [详细信息:Millipore Cat. no. MSBVS1210] Vacuum manifold [详细信息:Whatman Cat. no. 7705-0102]

实验步骤

 

1) Grow cells to desired level of confluency in a T75 flask.

2) Decant or aspirate the medium.

3) Add 2–3 ml fresh warm trypsin/EDTA solution. Transfer the flask to a 37°C incubator.

4) Wash with warm PBS. Aspirate.

5) After  5 minutes, tap the side of the flask, and examine the flask under a  microscope for lifting. If necessary, return the cells to the incubator  for an additional 5–10 minutes, with occasional tapping, until lifting  is complete.

6) Quickly quench the trypsin reaction by adding 5–6 ml complete cell culture medium.

7) Transfer the cells to sterile 15 ml conical tubes.

8) Pellet the cells by centrifugation at 300 x g for 7 minutes.

9) Decant the supernatant.

10)           Wash  the cells by pipetting 10 ml of medium into each conical tube and  resuspending the pellet. Collect the cells by centrifugation at 300 x g  for 7 minutes.

11)           Resuspend the washed cells in complete cell culture medium.

12)           Enumerate  cell density. A hemacytometer is recommended. For most applications,  the cell density should be adjusted to 25–100 x 104 cells/ml cell  culture medium. It is important to note that this value may require some  optimization for each specific application. Cell doubling time is an  important factor to be considered when adjusting cell density at the  beginning of an experiment.

13)           Plate  200 μl of cell culture (i.e., 50,000–200,000 cells) into the wells of  the sterile 96-well filter-bottom plate. Incubate the cells for 24 hours  at 37°C. The filter plate is designed to retain particles, while  permitting the flow of liquids from the bottom of the plate.

14)           Stimulate  the cells as desired. In the example presented below, Jurkat cells were  stimulated with 20 μg/ml anisomycin for 60 minutes at 37°C.

15)           At  the end of the stimulation, place the filter bottom plate on the vacuum  manifold and remove cell culture medium from the bottom of wells by  gentle aspiration.

16)           Wash  the cells by pipetting 200 μl ice-cold PBS into each well. Remove the  PBS from the bottom of the wells by gentle aspiration. Repeat two times  for a total of three washings.

17)           Pipette 30 μl protease-inhibitor-supplemented Cell Extraction Buffer into each well. Incubate the plate on ice for 30 minutes.

18)           Thoroughly  mix the contents of each well by pipetting up and down 5–6 times. A  multi-channel pipette is desirable for this application. At this point  in the procedure, the extracts are ready for analysis. Alternatively,  the extracts may be stored in the filter bottom plate at –20°C for  future analysis. Frozen plates should be thawed on ice in preparation of  completing the assays.

19)           Place the plate on an orbital shaker and mix for 1 minute.

20)           Prepare  the phosphoELISA™ kits. Sample Wells: Pipette 95 μl Standard Diluent  Buffer (included in the kits) into the wells of the phosphoELISA™ plates  designated for samples. Transfer 5 μl cell extract from the filter  plate into the sample wells of the plates. Place the plates on an  orbital shaker to thorougly mix the contents of the wells.

21)           Standard Wells: Prepare standards as indicated in the assay protocol and pipette into designated wells.

22)           Complete the phosphoELISA™ as directed by the assay protocol.

 


相关文章

毒蛇咬伤怎么办?这些检测方法来支招

2025年6月2日,一名游客在海南三亚疑似遭遇蛇咬伤,自行处理伤口后就医,仍因病情恶化最终不幸离世。6月3日,三亚市卫健委通报事件,引发公众广泛关注。这一事件令人痛惜,更提醒我们:毒蛇咬伤是对生命的直......

可复美陷入EGF风波:ELISA检测结果引争议,真相竟然是……

近日,关于可复美面膜产品被检测出违禁成分EGF(表皮生长因子)的事件引发了广泛关注。事件的发酵始于一位消费者在使用“可复美重组胶原蛋白修复敷料(G型)”后,面部出现疑似“馒化”现象。该消费者将产品送至......

大观霉素间接竞争性ELISA的建立及应用

大观霉素(spectinomycin,SPE)是一种氨基糖苷类广谱抗生素,是Mason于1961年首次从链霉菌中分离得到[1]。从结构上讲,氨基糖苷类抗生素分子中均含有氨基糖苷结构。而SPE(图1)是......

ELISA分析缺点

ELISA实验所有方法的缺点很明显:1、重复性不好;2、收自身抗体、嗜异性抗体等干扰,易出现假阳性;3、不论仪器和手工操作,干扰因素较多。影响最大的是温度和时间。1、直接法(directELISA)将......

酒中的草甘膦对人体健康的影响

最近,科学家们在啤酒和葡萄酒中发现了一种常见的杂草杀手—草甘膦。此前,它曾因其可能与癌症有关而引起争议。美国的研究人员测试了15种不同类型的啤酒和5种不同类型的葡萄酒,在20种饮料中的19种中发现了农......

ELISA的原理

ELISA的原理:(1)抗原或抗体能以物理性吸附于固相载体表面,可能是蛋白和聚苯乙烯表面间的疏水性部分相互吸附,并保持其免疫学活性;(2)抗原或抗体可通过共价键与酶连接形成酶结合物,而此种酶结合物仍能......

农业部公告第2593号牛支原体ELISA试剂盒等2种新兽药

根据《兽药管理条例》和《兽药注册办法》规定,经审查,批准中国农业科学院哈尔滨兽医研究所等6家单位申报的牛支原体ELISA抗体检测试剂盒等2种兽药产品为新兽药,核发《新兽药注册证书》,并发布产品试行规程......

ELISA实验中常见问题和注意事项

1临床标本的收集和保存用于ELISA测定的临床标本最为常用的是血清(浆)。本实验室用ELISA测定乙肝两对半,甲肝,戊肝,抗HIV的标本时,在处理和保存方面要考虑以下几个方面:1)要注意避免出现严重溶......

消除ELISA假阳性的改进措施

当你拿着化验单询问医生,他/她却告诉你需要做更多检查,你的心情想必更加焦虑。例行公事的医学检查好比针对潜在疾病的一场排雷竞赛。医生和患者们都如履薄冰,尽管执行了正确的检验,但其结果却错误地把不具备阳性......

ELISA实验的18条军规,你知道吗?

ELISA实验通用规则1、要保证移液枪的准确性,误差不能超过2%。可用水和电子天平进行确定。但最好有专业人员进行矫正。2、要配备20ul、50ul、100ul、1000ul和排枪各一支。吸取不同的液体......