RNAElectrophoresis
Electrophoresis through agarose or polyacrylamide gels is the standard way to separate, identify and purify nucleic acid fragments. The location of the nucleic acid within the gel can be determined by using the fluorescent intercalating dye ethidium bromide.Agarose gels have a smaller resolving power than polyacrylamide gels but a greater range of separation - from 200 bp to >50 kb using standard gels and ele......阅读全文
RNA-Electrophoresis
Electrophoresis through agarose or polyacrylamide gels is the standard way to separate, identify and purify nucleic acid fragments. The location of th
RNA-gel-electrophoresis
MaterialsDEPC H2ODEPC 0.1% (v/v)q.s. de-ioinized H2O37ºC x1 hr, or r.t. overnightAutoclave.(NaOAc, EDTA and ethidium bromide solutions should also be
RNA-gel-electrophoresis
实验概要RNA gel electrophoresis主要试剂DEPC H2ODEPC 0.1% (v/v)q.s. de-ioinized H2O37ºC x1 hr, or r.t. overnightAutoclave.(NaOAc, EDTA and ethidium bromide sol
Denaturing-Agarose-Gel-Electrophoresis-of-RNA
The overall quality of an RNA preparation may be assessed by electrophoresis on a denaturing agarose gel; this will also give some information about R
RNA-analysis-on-nondenaturing-agarose-gel-electrophoresis
1. The following gel electrophoresis conditions are recommended: - use 1X TAE buffer instead of 1X TBE - use agarose gel in the concentration o
RNA-analysis-on-nondenaturing-agarose-gel-electrophoresis
实验概要RNA analysis on non-denaturing agarose gel electrophoresis实验步骤1. The following gel electrophoresis conditions are recommended:- use 1X TAE buffer
Chromatin-Electrophoresis
Dr. William H. Heidcamp, Biology Department, Gustavus Adolphus College Exercise 10.4 - Chromatin Electrophoresis LEVEL II Materials 14 M Ure
Capillary-Electrophoresis
Capillary electrophoresis is a very sensitive analytical technique. Sample components are separated within a fused silica capillary using one of sever
electrophoresis-of-DNA
Agarose Gel Electroporesis of DNA Making the gel: 1. Place casting platform with well former sideways in gel stand where you wish to pour
Protein-Electrophoresis
DefinitionAmino acids, nucleotides, polypeptides, and other compounds in a colloidal state can be separated by the application of external voltages wh
DNA-Electrophoresis
What is Electrophoresis?Electrophoresis is a technique used in the laboratory that results in the separation of charged molecules. DNA is a negatively
Gel-Electrophoresis-of-DNA
What is Electrophoresis? Electrophoresis is a technique used in the laboratory that results in the separation of charged molecules. In this CyberLab
Pulse-Field-Electrophoresis
Manipulating and analyzing DNA are fundamentals in the field of molecular biology. Indeed, separating complex mixtures of DNA into different sized fra
Agarose-gel-electrophoresis
General ProcedureCast a gelPlace it in gel box in running bufferLoad samplesRun the gelImage the gelCasting Gels0.7% agarose gel with 1kbp ladder in U
Agarose-Gel-Electrophoresis
实验概要Separating nucleic acid fragments by agarose gel electrophoresis.实验原理 Agarose gel electrophoresis remains the most widely used technique for sep
Standard-neutral-agarose-electrophoresis
Standard neutral agarose electrophoresis Standard agarose gels can be prepared using either TBE or TAE running buffers. You will need: Either 10 x
Polyacrylamide-Gel-Electrophoresis-of-Oligonucleotides
1. Pour and polymerize a 20% polyacrylamide gel, no Urea. 2. Remove clamps. Rinse with water. Remove comb. Rinse top of gel well. 3. Insert comb te
Agarose-Gel-Electrophoresis-of-DNA
1) Dissolve 1 g of agarose in 100 ml of 1X TAE or TBE buffer (gives a 1% gel). See note for making LMP agarose gel. 2) Cast the gel with the comb in p
Blue-Native-Gel-Electrophoresis
Blue Native Gel ElectrophoresisStock solutions49.5%T, 3%C Acrylamide 24 g acrylamide, 0.75 g bisacrylamide / 50 ml H2O Store at RT3 x Gel buffer 150 m
ELECTROPHORESIS-OF-DNA-IN-AGAROSE-GELS
ELECTROPHORESIS OF DNA IN AGAROSE GELSA). AGAROSE CONCENTRATIONS: Use 0.8% agarose (w/v) for high molecular weight DNA fragments, and 1 - 1.2% f
Alkaline-agarose-gel-electrophoresis
Alkaline agarose gel electrophoresis (Sambrook et al., 1989)Alkaline agarose gels can be used to determine the size and quality of first and second st
ELECTROPHORESIS-OF-DNA-IN-POLYACRYLAMIDE-GELS
ELECTROPHORESIS OF DNA IN POLYACRYLAMIDE GELSGel SizesSmall: 165 x 130 mmMedium: 165 x 200 mmLarge: 165 x 260 mm5% Anal
SDS-Gel-Electrophoresis-of-Tubulin\MAPs
MaterialsStock Acrylamide: (30%T:0.8%C)30% by weight of acrylamide0.8% by weight of N,N'-bis-methylene acrylamideSeparation Gel (Final Concentrati
High-Resolution-Agarose-Gel-Electrophoresis
实验概要Agarose gel electrophoresis remains the most widely used technique for separating nucleic acid fragments due to its ease of use, non-toxicity, a
Denaturing-Gradient-Gel-Electrophoresis-(DGGE)
Purpose:Denaturing gradient gels are used to detect non-RFLP polymorphisms. The small (200-700 bp) genomic restriction fragments are run on a low to h
Lipoprotein-Analysis-Week-2:-Electrophoresis
Lipoprotein Analysis Week 2: Electrophoresis IntroductionSDS polyacrylamide gel electrophoresis (SDS PAGE) will be used to assess the purification pr
2D-Polyacrylamide-Gel-Electrophoresis
This method was successful in our lab using prostate tissue and for our specific objectives. Investigators must be aware that they will need to tailor
Electrophoresis-of-PCR-products-with-Sunrise-gel-apparatus
Electrophoresis of PCR products with Life Technologies Sunrise gel apparatusGel: In a 500 ml Pyrex® glass bottle, add:Agarose:3 gH2O270 mls10X TA30 ml
Top-10-Fun-Facts-for-DNA-Electrophoresis
Did you know:When preparing agarose for electrophoresis, it is best to sprinkle the agarose into room-temperature buffer, swirl, and let sit at least
Native-gel-electrophoresis(非变性电泳)
Native gel electrophoresis Under native PAGE conditions, polypeptides retain their higher-order structure and often retain enzymatic activity and inte